ICx Photonics offers a unique class of electrically pulsed, 
					high intensity infrared radiators for gas analysis, 
					spectroscopy, calibration and tactical identify friend or 
					foe (IFF) applications. These radiators feature a low 
					thermal-mass filament tailored for high emissivity. The 
					filament is fabricated using a patented process such that it 
					supplies bright IR power output while operating much cooler 
					than alternatives. This lower temperature operation reduces 
					the chance of igniting combustible gasses, improves power 
					efficiency, reduces heating of the optics and detectors, and 
					prevents illumination in the SWIR bands. These IR sources 
					are typically pulsed at rates from ½ to 10 Hz with several 
					hundred degrees of temperature modulation.
					
					For demonstration and system design, ICx Photonics provides 
					an Evaluation Kit that includes a power supply for your 
					region. The Evaluation Kit drive card produces a flat-topped 
					current pulse of adjustable amplitude, length, and frequency 
					that can be run with the pre-programmed settings, or can be 
					connected to a PC for user control via included Windows TM 
					software.
					
					The table below provides the specifications for each of our 
					broadband IR light sources. If you do not see a product that 
					meets your requirements, please contact us as we may be able 
					to provide a custom solution that meets your needs. Also, we 
					have a line of narrowband IR light sources which use MEMS 
					technology. Because of their spectral tuning, these are 
					extremely efficient devices suitable for battery powered 
					applications.
SPECIFICATIONS, PART NUMBERS, and WINDOW 
					OPTIONS:
					
						
							
								|  | Parabola | TO-8 | Mulit-Element | TO-5 | TO-46 | 
							
								|  | 
								 | 
								 | 
								 | 
								 | 
								 | 
							
								| 
									2 Element | 
									4 Element | 
							
								| Windowless | reflectIR-P1N | NL8LNC | NL82LNC | NL84LNC | NM5LNC | NL46LNX | 
							
								| Sapphire 2 to 5.25µm
 | relfectIR-P1S | NM8ASC | NM82ASC | NM84ASC | NM5NSC | N/A | 
							
								| Germanium 7 to 12µm
 | N/A | N/A | N/A | N/A | NL5NGC | N/A | 
							
								| Calcium Flouride
 2 to 9.5µm
 | reflectIR-P1C | NL8ACC | NL82ACC | NL84ACC | NL5NCC | N/A | 
							
								| Rated Temperature
 | 750°C | 
							
								| Minimum Resistance
 | 1.4 Ohms | 2.8 Ω | 1.3 Ω per Element | 2.5 Ω | 0.4 Ω | 
							
								| Maximum Resistance
 | 2.0Ohms | 4.5 Ω | 1.7 Ω per Element | 3.7 Ω | 1.0 Ω | 
							
								| Maximum Input
 Power
 | 1.7 W | 2.2 W | 1.6 W per Element | 2 W | 1.1 W | 
							
								| Modulation Speed
 | Constant - 30hz; 100% modulation <5hz | 
							
								| Output Radiation
 Pattern*
 | 
								 | 
								 | 
					
					
					
 PULSED OPERATION
					Although capable of running at duty cycles of up to 100 % 
					(DC) most users run the filaments with duty cycles of less 
					than 50%. Square-waveform constant current or constant 
					voltage drive schemes are the simplest and most cost 
					effective means of powering the sources. For constant 
					current drives, the power delivered to the source goes as I2 
					x R. As the source heats up, its resistance increases 
					slightly, causing the power delivered to the source to 
					increase during the “ON” portion of a pulse. For constant 
					voltage drives, delivered power goes as V2/R; therefore the 
					power delivered to the source tends to decrease slightly 
					during the length of a pulse. Other drive schemes can also 
					be employed; constant power or DC for example.
					
					Owing to the extremely low thermal mass of pulsIR emitters, 
					shot-to-shot stability is directly related to drive circuit 
					stability. Variations in drive pulses will translate into 
					variations in output. To determine this we used a liquid 
					nitrogen cooled InSb detector available in our laboratory 
					for detecting energy in the 2-5 μm range. The pulsIR source 
					was driven with a constant-voltage drive circuit that 
					ensures pulse-to-pulse repeatability (standard deviation) of 
					5.3x10-4. Measurements of the InSb detector reading from 16 
					seconds of 10 Hz operation was measured to have a comparable 
					standard deviation of 6.8x10-4.
					SOURCE LIFETIME
					The following graph shows the results (to date) from an 
					ongoing extended life test experiment using an ICx Photonics 
					NM8ASC source. The source is being driven by a constant 
					current drive board at 1 Hz, 30% duty cycle at an 
					approximate temperature of 650°C. Two pyroelectric etectors 
					are monitoring the source output at two distinct 
					wavelengths. In the following chart, the circles show the 
					source output at 4.29 microns (CO2) while the diamonds show 
					the output at 3.9 microns (reference). The detectors are 
					mounted about four inches from the front face of the source 
					and a dry nitrogen purge is used to prevent water vapor and 
					carbon dioxide in the lab air from affecting the 
					measurement. The temperature in the lab is not very well 
					controlled however, and much of the variation (specifically 
					the bump at ~2000 hours) is due to room temperature swings.
					
					
					The definition of failure, and thus the definition of 
					lifetime, is very subjective as each system has unique 
					sensitivity to drift (largely related to the A/D bandwidth). 
					We have encountered several applications which define 
					failure as >15% drift from the original power level, so we 
					will adopt this definition for the purposes of this 
					computation. The graph below shows that the median signal 
					level from the 3.9 and 4.29 μm detectors is roughly 4 volts; 
					the linear regression fits to the raw data indicate that 
					both of these signals are decreasing at a rate of 1x10-5 
					volts/hour. With our assumed signal drift tolerance of 15% 
					and 4 volt signal level, we require a 0.6 volt signal change 
					to signal failure of the light source [0.15 x 4]. With our 
					measured rate of change being 1x10-5 volts/hour it will take 
					approximately 7 years of continuous operation to obtain a 
					15% signal change [(0.6v)/(1x10-5v/hr)/(8760hrs/yr)]. Since 
					many systems utilize the ratio of the gas measurement to a 
					reference, they are sensitive not to signal changes, but to 
					change in the ratio of the two signals. With a measured 
					slope of 1x10-6 volts/hour and a 0.75 volt signal the same 
					computation yields a lifetime of 13 years. Since all of the 
					known filament degradation mechanisms are temperature 
					dependent, the time to 15% failure is strongly dependent 
					upon operating temperature or electrical power applied. 
					Therefore, caution should be used in extrapolating these 
					results to your application.